Secondary Emission from Metallic Surfaces
Geant4 Version: 4.11.3 (OpenTOPAS 4.0)
Operating System: Linux
Dear Geant4 Community,
I am currently running simulations of a Faraday cup, where I want to check the influence of the electric field generated by a charged guard ring on the charge collection efficiency. I am simulating the interaction of 70 and 150 MeV clinical proton beams with the detector, and scoring the collected charge in the cup with respect to the incident charge.
To adequately model surface emission, I have applied the MicroElec physics list on 10nm thick volumes corresponding to the metallic surfaces of the detector. Range cuts have been fixed to 1nm and the low edge for production to 0.1 eV. The lowest electron energy is set to 0 eV.
The following plot showcases my results so far:
As shown, in the range [-100 V, 100 V], the simulation matches pretty well with the measurements. However, for lower or higher field settings, there is a mismatch. In particular, for negative potentials the higher signal simulated suggest that there is some missing excess negative charge in the absorber, and the difference for positive potentials suggest that not enough negative charge is being removed from the absorber (due to backscattering).
I attach below the TOPAS parameter file I am using to run one of these simulations (note that some implementations, such as the microelec physics list and mapped electric fields, are custom and not included in TOPAS release versions).
Do you have any suggestion on how to enhance secondary emission to better match the measurements, if that is the problem?
Thank you very much.
Marc
# File: input.tps
b:Ge/QuitIfOverlapDetected = "true"
b:Ts/PauseBeforeQuit = "false"
i:Ts/RandomSeed = 42
i:Ts/NumberOfThreads = 0
d:Ph/Default/ForRegion/highprecision/CutForElectron = 1e-06 mm
d:Ph/Default/ForRegion/highprecision/CutForPositron = 1e-06 mm
d:Ph/Default/ForRegion/highprecision/CutForProton = 1e-06 mm
d:Ph/Default/ForRegion/highprecision/CutForGamma = 1e-06 mm
d:Ph/Default/ForRegion/highprecision/CutForAlpha = 1e-06 mm
b:Ph/Default/PIXE = "true"
b:Ph/Default/Auger = "true"
b:Ph/Default/AugerCascade = "true"
b:Ph/Default/Fluorescence = "true"
i:Ph/Default/EMBins = 220
d:Ph/Default/EMRangeMin = 1.0 eV
d:Ph/Default/LowestElectronEnergy = 0.1 eV
d:Ph/Default/SetProductionCutLowerEdge = 0.1 eV
d:Ph/Default/ForRegion/microelec/CutForElectron = 1e-06 mm
d:Ph/Default/ForRegion/microelec/CutForPositron = 1e-06 mm
d:Ph/Default/ForRegion/microelec/CutForProton = 1e-06 mm
d:Ph/Default/ForRegion/microelec/CutForGamma = 1e-06 mm
d:Ph/Default/ForRegion/microelec/CutForAlpha = 1e-06 mm
sv:Ph/Default/Modules = 7 "g4em-microelec" "g4em-standard_opt4" "g4h-phy_QGSP_BIC_HP" "g4decay" "g4ion-binarycascade" "g4h-elastic_HP" "g4stopping"
b:Ts/UseQt = "false"
s:Ge/World/Material = "G4_AIR"
d:Ge/World/HLX = 0.25 m
d:Ge/World/HLY = 0.25 m
d:Ge/World/HLZ = 0.5 m
sc:Ge/World/DrawingStyle = "WireFrame"
sc:Ge/World/Color = "white"
s:Ge/FieldContainer/Type = "TsBox"
s:Ge/FieldContainer/Parent = "World"
s:Ge/FieldContainer/Material = Ge/World/Material
d:Ge/FieldContainer/HLX = Ge/World/HLX m
d:Ge/FieldContainer/HLY = Ge/World/HLY m
d:Ge/FieldContainer/HLZ = Ge/World/HLZ m
d:Ge/FieldContainer/TransX = 0 mm
d:Ge/FieldContainer/TransY = 0 mm
d:Ge/FieldContainer/TransZ = 0 mm
d:Ge/FieldContainer/RotX = 0 deg
d:Ge/FieldContainer/RotY = 0 deg
d:Ge/FieldContainer/RotZ = 0 deg
sc:Ge/FieldContainer/Color = "grey220"
sc:Ge/FieldContainer/DrawingStyle = "WireFrame"
s:Ge/VacuumWindowGroup/Type = "Group"
s:Ge/VacuumWindowGroup/Parent = "FieldContainer"
s:Ge/VacuumWindowGroup/Material = "G4_Al"
d:Ge/VacuumWindowGroup/TransX = 0 mm
d:Ge/VacuumWindowGroup/TransY = 0 mm
d:Ge/VacuumWindowGroup/TransZ = -0.05 mm
d:Ge/VacuumWindowGroup/RotX = 0 deg
d:Ge/VacuumWindowGroup/RotY = 0 deg
d:Ge/VacuumWindowGroup/RotZ = 0 deg
s:Ge/VacuumWindow/Type = "TsCylinder"
s:Ge/VacuumWindow/Parent = "VacuumWindowGroup"
s:Ge/VacuumWindow/Material = "G4_Al"
s:Ge/VacuumWindow/AssignToRegionNamed = "highprecision"
d:Ge/VacuumWindow/RMax = 71.5 mm
d:Ge/VacuumWindow/HL = 0.049975000000000006 mm
d:Ge/VacuumWindow/TransX = 0 mm
d:Ge/VacuumWindow/TransY = 0 mm
d:Ge/VacuumWindow/TransZ = 2.5e-05 mm
d:Ge/VacuumWindow/RotX = 0 deg
d:Ge/VacuumWindow/RotY = 0 deg
d:Ge/VacuumWindow/RotZ = 0 deg
sc:Ge/VacuumWindow/Color = "grey220"
sc:Ge/VacuumWindow/DrawingStyle = "Solid"
s:Ge/VacuumWindowMicroelecLayer/Type = "TsCylinder"
s:Ge/VacuumWindowMicroelecLayer/Parent = "VacuumWindowGroup"
s:Ge/VacuumWindowMicroelecLayer/Material = "G4_Al"
s:Ge/VacuumWindowMicroelecLayer/AssignToRegionNamed = "microelec"
d:Ge/VacuumWindowMicroelecLayer/RMax = 71.5 mm
d:Ge/VacuumWindowMicroelecLayer/HL = 2.5e-05 mm
d:Ge/VacuumWindowMicroelecLayer/TransX = 0 mm
d:Ge/VacuumWindowMicroelecLayer/TransY = 0 mm
d:Ge/VacuumWindowMicroelecLayer/TransZ = -0.049975000000000006 mm
d:Ge/VacuumWindowMicroelecLayer/RotX = 0 deg
d:Ge/VacuumWindowMicroelecLayer/RotY = 0 deg
d:Ge/VacuumWindowMicroelecLayer/RotZ = 0 deg
sc:Ge/VacuumWindowMicroelecLayer/Color = "grey220"
sc:Ge/VacuumWindowMicroelecLayer/DrawingStyle = "Solid"
s:Ge/Vacuum/Type = "TsCylinder"
s:Ge/Vacuum/Parent = "FieldContainer"
s:Ge/Vacuum/Material = "FCVacuum"
s:Ge/Vacuum/AssignToRegionNamed = "highprecision"
sv:Ma/FCVacuum/Components = 4 "Carbon" "Nitrogen" "Oxygen" "Argon"
uv:Ma/FCVacuum/Fractions = 4 0.000124 0.755267 0.231781 0.012828
s:Ma/FCVacuum/State = "Gas"
d:Ma/FCVacuum/Density = 1.19e-10 g/cm3
d:Ma/FCVacuum/Temperature = 293 K
d:Ma/FCVacuum/Pressure = 0.01 Pa
d:Ge/Vacuum/RMax = 74.0 mm
d:Ge/Vacuum/HL = 196.0 mm
d:Ge/Vacuum/TransX = 0 mm
d:Ge/Vacuum/TransY = 0 mm
d:Ge/Vacuum/TransZ = -196.1 mm
d:Ge/Vacuum/RotX = 0 deg
d:Ge/Vacuum/RotY = 0 deg
d:Ge/Vacuum/RotZ = 0 deg
sc:Ge/Vacuum/Color = "lightblue"
sc:Ge/Vacuum/DrawingStyle = "WireFrame"
s:Ge/Mantle/Type = "TsCylinder"
s:Ge/Mantle/Parent = "FieldContainer"
s:Ge/Mantle/Material = "Steel"
s:Ge/Mantle/AssignToRegionNamed = "highprecision"
d:Ge/Mantle/RMax = 79.0 mm
d:Ge/Mantle/RMin = 74.0 mm
d:Ge/Mantle/HL = 196.05 mm
d:Ge/Mantle/TransX = 0 mm
d:Ge/Mantle/TransY = 0 mm
d:Ge/Mantle/TransZ = -196.05 mm
d:Ge/Mantle/RotX = 0 deg
d:Ge/Mantle/RotY = 0 deg
d:Ge/Mantle/RotZ = 0 deg
sc:Ge/Mantle/Color = "Silver"
sc:Ge/Mantle/DrawingStyle = "FullWireFrame"
s:Ge/GuardRing/Type = "Group"
s:Ge/GuardRing/Parent = "Vacuum"
d:Ge/GuardRing/TransX = 0 mm
d:Ge/GuardRing/TransY = 0 mm
d:Ge/GuardRing/TransZ = 86.5 mm
d:Ge/GuardRing/RotX = 0 deg
d:Ge/GuardRing/RotY = 0 deg
d:Ge/GuardRing/RotZ = 0 deg
s:Ge/GuardRingBulk/Type = "TsCylinder"
s:Ge/GuardRingBulk/Parent = "GuardRing"
s:Ge/GuardRingBulk/Material = "Aluminum"
s:Ge/GuardRingBulk/AssignToRegionNamed = "highprecision"
d:Ge/GuardRingBulk/RMax = 67.19999 mm
d:Ge/GuardRingBulk/RMin = 63.50001 mm
d:Ge/GuardRingBulk/HL = 77.49999 mm
d:Ge/GuardRingBulk/TransX = 0 mm
d:Ge/GuardRingBulk/TransY = 0 mm
d:Ge/GuardRingBulk/TransZ = 0 mm
d:Ge/GuardRingBulk/RotX = 0 deg
d:Ge/GuardRingBulk/RotY = 0 deg
d:Ge/GuardRingBulk/RotZ = 0 deg
sc:Ge/GuardRingBulk/Color = "silver"
sc:Ge/GuardRingBulk/DrawingStyle = "Solid"
s:Ge/GuardRingInnerLayer/Type = "TsCylinder"
s:Ge/GuardRingInnerLayer/Parent = "GuardRing"
s:Ge/GuardRingInnerLayer/Material = "G4_Al"
s:Ge/GuardRingInnerLayer/AssignToRegionNamed = "microelec"
d:Ge/GuardRingInnerLayer/RMax = 63.50001 mm
d:Ge/GuardRingInnerLayer/RMin = 63.5 mm
d:Ge/GuardRingInnerLayer/HL = 77.49999 mm
d:Ge/GuardRingInnerLayer/TransX = 0 mm
d:Ge/GuardRingInnerLayer/TransY = 0 mm
d:Ge/GuardRingInnerLayer/TransZ = 0 mm
d:Ge/GuardRingInnerLayer/RotX = 0 deg
d:Ge/GuardRingInnerLayer/RotY = 0 deg
d:Ge/GuardRingInnerLayer/RotZ = 0 deg
sc:Ge/GuardRingInnerLayer/Color = "maroon"
sc:Ge/GuardRingInnerLayer/DrawingStyle = "Solid"
s:Ge/GuardRingOuterLayer/Type = "TsCylinder"
s:Ge/GuardRingOuterLayer/Parent = "GuardRing"
s:Ge/GuardRingOuterLayer/Material = "G4_Al"
s:Ge/GuardRingOuterLayer/AssignToRegionNamed = "microelec"
d:Ge/GuardRingOuterLayer/RMax = 67.2 mm
d:Ge/GuardRingOuterLayer/RMin = 67.19999 mm
d:Ge/GuardRingOuterLayer/HL = 77.49999 mm
d:Ge/GuardRingOuterLayer/TransX = 0 mm
d:Ge/GuardRingOuterLayer/TransY = 0 mm
d:Ge/GuardRingOuterLayer/TransZ = 0 mm
d:Ge/GuardRingOuterLayer/RotX = 0 deg
d:Ge/GuardRingOuterLayer/RotY = 0 deg
d:Ge/GuardRingOuterLayer/RotZ = 0 deg
sc:Ge/GuardRingOuterLayer/Color = "maroon"
sc:Ge/GuardRingOuterLayer/DrawingStyle = "Solid"
s:Ge/GuardRingTopLayer/Type = "TsCylinder"
s:Ge/GuardRingTopLayer/Parent = "GuardRing"
s:Ge/GuardRingTopLayer/Material = "G4_Al"
s:Ge/GuardRingTopLayer/AssignToRegionNamed = "microelec"
d:Ge/GuardRingTopLayer/RMax = 67.2 mm
d:Ge/GuardRingTopLayer/RMin = 63.5 mm
d:Ge/GuardRingTopLayer/HL = 1e-05 mm
d:Ge/GuardRingTopLayer/TransX = 0 mm
d:Ge/GuardRingTopLayer/TransY = 0 mm
d:Ge/GuardRingTopLayer/TransZ = 77.500005 mm
d:Ge/GuardRingTopLayer/RotX = 0 deg
d:Ge/GuardRingTopLayer/RotY = 0 deg
d:Ge/GuardRingTopLayer/RotZ = 0 deg
sc:Ge/GuardRingTopLayer/Color = "maroon"
sc:Ge/GuardRingTopLayer/DrawingStyle = "Solid"
s:Ge/GuardRingBottomLayer/Type = "TsCylinder"
s:Ge/GuardRingBottomLayer/Parent = "GuardRing"
s:Ge/GuardRingBottomLayer/Material = "G4_Al"
s:Ge/GuardRingBottomLayer/AssignToRegionNamed = "microelec"
d:Ge/GuardRingBottomLayer/RMax = 67.2 mm
d:Ge/GuardRingBottomLayer/RMin = 63.5 mm
d:Ge/GuardRingBottomLayer/HL = 1e-05 mm
d:Ge/GuardRingBottomLayer/TransX = 0 mm
d:Ge/GuardRingBottomLayer/TransY = 0 mm
d:Ge/GuardRingBottomLayer/TransZ = -77.500005 mm
d:Ge/GuardRingBottomLayer/RotX = 0 deg
d:Ge/GuardRingBottomLayer/RotY = 0 deg
d:Ge/GuardRingBottomLayer/RotZ = 0 deg
sc:Ge/GuardRingBottomLayer/Color = "maroon"
sc:Ge/GuardRingBottomLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockWalls/Type = "Group"
s:Ge/AbsorberBlockWalls/Parent = "Vacuum"
d:Ge/AbsorberBlockWalls/TransX = 0 mm
d:Ge/AbsorberBlockWalls/TransY = 0 mm
d:Ge/AbsorberBlockWalls/TransZ = -46.0 mm
d:Ge/AbsorberBlockWalls/RotX = 0 deg
d:Ge/AbsorberBlockWalls/RotY = 0 deg
d:Ge/AbsorberBlockWalls/RotZ = 0 deg
s:Ge/AbsorberBlockWallsBulk/Type = "TsCylinder"
s:Ge/AbsorberBlockWallsBulk/Parent = "AbsorberBlockWalls"
s:Ge/AbsorberBlockWallsBulk/Material = "Brass"
s:Ge/AbsorberBlockWallsBulk/AssignToRegionNamed = "highprecision"
d:Ge/AbsorberBlockWallsBulk/RMax = 67.19995 mm
d:Ge/AbsorberBlockWallsBulk/RMin = 63.50005 mm
d:Ge/AbsorberBlockWallsBulk/HL = 49.999975 mm
d:Ge/AbsorberBlockWallsBulk/TransX = 0 mm
d:Ge/AbsorberBlockWallsBulk/TransY = 0 mm
d:Ge/AbsorberBlockWallsBulk/TransZ = -2.5e-05 mm
d:Ge/AbsorberBlockWallsBulk/RotX = 0 deg
d:Ge/AbsorberBlockWallsBulk/RotY = 0 deg
d:Ge/AbsorberBlockWallsBulk/RotZ = 0 deg
sc:Ge/AbsorberBlockWallsBulk/Color = "orange"
sc:Ge/AbsorberBlockWallsBulk/DrawingStyle = "Solid"
s:Ge/AbsorberBlockWallsInnerLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockWallsInnerLayer/Parent = "AbsorberBlockWalls"
s:Ge/AbsorberBlockWallsInnerLayer/Material = "G4_Cu"
s:Ge/AbsorberBlockWallsInnerLayer/AssignToRegionNamed = "microelec"
d:Ge/AbsorberBlockWallsInnerLayer/RMax = 63.50005 mm
d:Ge/AbsorberBlockWallsInnerLayer/RMin = 63.5 mm
d:Ge/AbsorberBlockWallsInnerLayer/HL = 49.999975 mm
d:Ge/AbsorberBlockWallsInnerLayer/TransX = 0 mm
d:Ge/AbsorberBlockWallsInnerLayer/TransY = 0 mm
d:Ge/AbsorberBlockWallsInnerLayer/TransZ = -2.5e-05 mm
d:Ge/AbsorberBlockWallsInnerLayer/RotX = 0 deg
d:Ge/AbsorberBlockWallsInnerLayer/RotY = 0 deg
d:Ge/AbsorberBlockWallsInnerLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockWallsInnerLayer/Color = "brown"
sc:Ge/AbsorberBlockWallsInnerLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockWallsOuterLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockWallsOuterLayer/Parent = "AbsorberBlockWalls"
s:Ge/AbsorberBlockWallsOuterLayer/Material = "G4_Cu"
s:Ge/AbsorberBlockWallsOuterLayer/AssignToRegionNamed = "microelec"
d:Ge/AbsorberBlockWallsOuterLayer/RMax = 67.2 mm
d:Ge/AbsorberBlockWallsOuterLayer/RMin = 67.19995 mm
d:Ge/AbsorberBlockWallsOuterLayer/HL = 49.999975 mm
d:Ge/AbsorberBlockWallsOuterLayer/TransX = 0 mm
d:Ge/AbsorberBlockWallsOuterLayer/TransY = 0 mm
d:Ge/AbsorberBlockWallsOuterLayer/TransZ = -2.5e-05 mm
d:Ge/AbsorberBlockWallsOuterLayer/RotX = 0 deg
d:Ge/AbsorberBlockWallsOuterLayer/RotY = 0 deg
d:Ge/AbsorberBlockWallsOuterLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockWallsOuterLayer/Color = "brown"
sc:Ge/AbsorberBlockWallsOuterLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockWallsTopLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockWallsTopLayer/Parent = "AbsorberBlockWalls"
s:Ge/AbsorberBlockWallsTopLayer/Material = "G4_Cu"
s:Ge/AbsorberBlockWallsTopLayer/AssignToRegionNamed = "microelec"
d:Ge/AbsorberBlockWallsTopLayer/RMax = 67.2 mm
d:Ge/AbsorberBlockWallsTopLayer/RMin = 63.5 mm
d:Ge/AbsorberBlockWallsTopLayer/HL = 2.5e-05 mm
d:Ge/AbsorberBlockWallsTopLayer/TransX = 0 mm
d:Ge/AbsorberBlockWallsTopLayer/TransY = 0 mm
d:Ge/AbsorberBlockWallsTopLayer/TransZ = 49.999975 mm
d:Ge/AbsorberBlockWallsTopLayer/RotX = 0 deg
d:Ge/AbsorberBlockWallsTopLayer/RotY = 0 deg
d:Ge/AbsorberBlockWallsTopLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockWallsTopLayer/Color = "brown"
sc:Ge/AbsorberBlockWallsTopLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockSolidCollector/Type = "Group"
s:Ge/AbsorberBlockSolidCollector/Parent = "Vacuum"
d:Ge/AbsorberBlockSolidCollector/TransX = 0 mm
d:Ge/AbsorberBlockSolidCollector/TransY = 0 mm
d:Ge/AbsorberBlockSolidCollector/TransZ = -146.00000000000003 mm
d:Ge/AbsorberBlockSolidCollector/RotX = 0 deg
d:Ge/AbsorberBlockSolidCollector/RotY = 0 deg
d:Ge/AbsorberBlockSolidCollector/RotZ = 0 deg
s:Ge/AbsorberBlockSolidCollectorTopLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockSolidCollectorTopLayer/Parent = "AbsorberBlockSolidCollector"
s:Ge/AbsorberBlockSolidCollectorTopLayer/Material = "G4_Cu"
s:Ge/AbsorberBlockSolidCollectorTopLayer/AssignToRegionNamed = "microelec"
d:Ge/AbsorberBlockSolidCollectorTopLayer/RMax = 67.2 mm
d:Ge/AbsorberBlockSolidCollectorTopLayer/HL = 2.5e-05 mm
d:Ge/AbsorberBlockSolidCollectorTopLayer/TransX = 0 mm
d:Ge/AbsorberBlockSolidCollectorTopLayer/TransY = 0 mm
d:Ge/AbsorberBlockSolidCollectorTopLayer/TransZ = 49.999975 mm
d:Ge/AbsorberBlockSolidCollectorTopLayer/RotX = 0 deg
d:Ge/AbsorberBlockSolidCollectorTopLayer/RotY = 0 deg
d:Ge/AbsorberBlockSolidCollectorTopLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockSolidCollectorTopLayer/Color = "brown"
sc:Ge/AbsorberBlockSolidCollectorTopLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockSolidCollectorBulk/Type = "TsCylinder"
s:Ge/AbsorberBlockSolidCollectorBulk/Parent = "AbsorberBlockSolidCollector"
s:Ge/AbsorberBlockSolidCollectorBulk/Material = "Brass"
s:Ge/AbsorberBlockSolidCollectorBulk/AssignedToRegionNamed = "highprecision"
d:Ge/AbsorberBlockSolidCollectorBulk/RMax = 67.19995 mm
d:Ge/AbsorberBlockSolidCollectorBulk/HL = 49.99995 mm
d:Ge/AbsorberBlockSolidCollectorBulk/TransX = 0 mm
d:Ge/AbsorberBlockSolidCollectorBulk/TransY = 0 mm
d:Ge/AbsorberBlockSolidCollectorBulk/TransZ = 0 mm
d:Ge/AbsorberBlockSolidCollectorBulk/RotX = 0 deg
d:Ge/AbsorberBlockSolidCollectorBulk/RotY = 0 deg
d:Ge/AbsorberBlockSolidCollectorBulk/RotZ = 0 deg
sc:Ge/AbsorberBlockSolidCollectorBulk/Color = "orange"
sc:Ge/AbsorberBlockSolidCollectorBulk/DrawingStyle = "Solid"
s:Ge/AbsorberBlockSolidCollectorOuterLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockSolidCollectorOuterLayer/Parent = "AbsorberBlockSolidCollector"
s:Ge/AbsorberBlockSolidCollectorOuterLayer/Material = "Brass"
s:Ge/AbsorberBlockSolidCollectorOuterLayer/AssignedToRegionNamed = "highprecision"
d:Ge/AbsorberBlockSolidCollectorOuterLayer/RMax = 67.2 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/RMin = 67.19995 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/HL = 49.99995 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/TransX = 0 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/TransY = 0 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/TransZ = 0 mm
d:Ge/AbsorberBlockSolidCollectorOuterLayer/RotX = 0 deg
d:Ge/AbsorberBlockSolidCollectorOuterLayer/RotY = 0 deg
d:Ge/AbsorberBlockSolidCollectorOuterLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockSolidCollectorOuterLayer/Color = "brown"
sc:Ge/AbsorberBlockSolidCollectorOuterLayer/DrawingStyle = "Solid"
s:Ge/AbsorberBlockSolidCollectorBottomLayer/Type = "TsCylinder"
s:Ge/AbsorberBlockSolidCollectorBottomLayer/Parent = "AbsorberBlockSolidCollector"
s:Ge/AbsorberBlockSolidCollectorBottomLayer/Material = "G4_Cu"
s:Ge/AbsorberBlockSolidCollectorBottomLayer/AssignToRegionNamed = "microelec"
d:Ge/AbsorberBlockSolidCollectorBottomLayer/RMax = 67.2 mm
d:Ge/AbsorberBlockSolidCollectorBottomLayer/HL = 2.5e-05 mm
d:Ge/AbsorberBlockSolidCollectorBottomLayer/TransX = 0 mm
d:Ge/AbsorberBlockSolidCollectorBottomLayer/TransY = 0 mm
d:Ge/AbsorberBlockSolidCollectorBottomLayer/TransZ = -49.999975 mm
d:Ge/AbsorberBlockSolidCollectorBottomLayer/RotX = 0 deg
d:Ge/AbsorberBlockSolidCollectorBottomLayer/RotY = 0 deg
d:Ge/AbsorberBlockSolidCollectorBottomLayer/RotZ = 0 deg
sc:Ge/AbsorberBlockSolidCollectorBottomLayer/Color = "brown"
sc:Ge/AbsorberBlockSolidCollectorBottomLayer/DrawingStyle = "Solid"
s:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/Component = "AbsorberBlockSolidCollectorTopLayer"
sv:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/OutputFile = "AbsorberBlockSolidCollectorTopLayer__charge_all"
s:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockSolidCollectorTopLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockSolidCollectorBulk__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockSolidCollectorBulk__charge_all/Component = "AbsorberBlockSolidCollectorBulk"
sv:Sc/AbsorberBlockSolidCollectorBulk__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockSolidCollectorBulk__charge_all/OutputFile = "AbsorberBlockSolidCollectorBulk__charge_all"
s:Sc/AbsorberBlockSolidCollectorBulk__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockSolidCollectorBulk__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/Component = "AbsorberBlockSolidCollectorOuterLayer"
sv:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/OutputFile = "AbsorberBlockSolidCollectorOuterLayer__charge_all"
s:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockSolidCollectorOuterLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/Component = "AbsorberBlockSolidCollectorBottomLayer"
sv:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/OutputFile = "AbsorberBlockSolidCollectorBottomLayer__charge_all"
s:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockSolidCollectorBottomLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockWallsBulk__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockWallsBulk__charge_all/Component = "AbsorberBlockWallsBulk"
sv:Sc/AbsorberBlockWallsBulk__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockWallsBulk__charge_all/OutputFile = "AbsorberBlockWallsBulk__charge_all"
s:Sc/AbsorberBlockWallsBulk__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockWallsBulk__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockWallsInnerLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockWallsInnerLayer__charge_all/Component = "AbsorberBlockWallsInnerLayer"
sv:Sc/AbsorberBlockWallsInnerLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockWallsInnerLayer__charge_all/OutputFile = "AbsorberBlockWallsInnerLayer__charge_all"
s:Sc/AbsorberBlockWallsInnerLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockWallsInnerLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockWallsOuterLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockWallsOuterLayer__charge_all/Component = "AbsorberBlockWallsOuterLayer"
sv:Sc/AbsorberBlockWallsOuterLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockWallsOuterLayer__charge_all/OutputFile = "AbsorberBlockWallsOuterLayer__charge_all"
s:Sc/AbsorberBlockWallsOuterLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockWallsOuterLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockWallsTopLayer__charge_all/Quantity = "Charge"
s:Sc/AbsorberBlockWallsTopLayer__charge_all/Component = "AbsorberBlockWallsTopLayer"
sv:Sc/AbsorberBlockWallsTopLayer__charge_all/Report = 2 "sum" "variance"
s:Sc/AbsorberBlockWallsTopLayer__charge_all/OutputFile = "AbsorberBlockWallsTopLayer__charge_all"
s:Sc/AbsorberBlockWallsTopLayer__charge_all/OutputFileType = "csv"
s:Sc/AbsorberBlockWallsTopLayer__charge_all/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/Quantity = "PhaseSpace"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/Surface = "VacuumWindow/zplussurface"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/OutputType = "Binary"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/OutputFile = "VacuumWindow@zplussurface__ps_proton_in"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/OutputFileType = "Binary"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/VacuumWindow@zplussurface__ps_proton_in/OnlyIncludeParticlesGoing = "in"
sv:Sc/VacuumWindow@zplussurface__ps_proton_in/OnlyIncludeParticlesNamed = 1 "proton"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/Quantity = "PhaseSpace"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/Surface = "VacuumWindowMicroelecLayer/zminussurface"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/OutputType = "Binary"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/OutputFile = "VacuumWindowMicroelecLayer@zminussurface__ps_all_out"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/OutputFileType = "Binary"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/VacuumWindowMicroelecLayer@zminussurface__ps_all_out/OnlyIncludeParticlesGoing = "in"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/Quantity = "PhaseSpace"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/Surface = "AbsorberBlockSolidCollectorTopLayer/zplussurface"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/OutputType = "Binary"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/OutputFile = "AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/OutputFileType = "Binary"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockSolidCollectorTopLayer@zplussurface__ps_all_out/OnlyIncludeParticlesGoing = "out"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/Quantity = "PhaseSpace"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/Surface = "AbsorberBlockSolidCollectorBottomLayer/zminussurface"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/OutputType = "Binary"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/OutputFile = "AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/OutputFileType = "Binary"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/IfOutputFileAlreadyExists = "Overwrite"
s:Sc/AbsorberBlockSolidCollectorBottomLayer@zminussurface__ps_all_out/OnlyIncludeParticlesGoing = "out"
s:Ge/FieldContainer/Field = "ElectricFieldMap"
s:Ge/FieldContainer/ElectricField3DTable = "/data/user/balles_m/FaradayCup/data/input/field_tables/electric/csv/EField_-1000V.csv"
s:Ge/SourceOrigin/Parent = "FieldContainer"
s:Ge/SourceOrigin/Type = "Group"
d:Ge/SourceOrigin/TransX = 0.0 mm
d:Ge/SourceOrigin/TransY = 0.0 mm
d:Ge/SourceOrigin/TransZ = 478.0 mm
d:Ge/SourceOrigin/RotX = 0.0 deg
d:Ge/SourceOrigin/RotY = 180.0 deg
d:Ge/SourceOrigin/RotZ = 0.0 deg
s:So/Source/Component = "SourceOrigin"
s:So/Source/BeamParticle = "proton"
s:So/Source/Type = "emittance"
s:So/Source/Distribution = "bigaussian"
d:So/Source/BeamEnergy = 70.939 MeV
u:So/Source/BeamEnergySpread = 0.777402
d:So/Source/SigmaX = 4.71598 mm
u:So/Source/SigmaXPrime = 0.0082
d:So/Source/SigmaY = 3.91972 mm
u:So/Source/SigmaYPrime = 0.0079
u:So/Source/CorrelationX = -0.095231
u:So/Source/CorrelationY = 0.118743
i:So/Source/NumberOfHistoriesInRun = 1000000