Silicon K-fluorescence yield: 0.93% (G4EMLOW) vs 4.5% (literature)


• Geant4 version: 11.3.1 (patch-01)
• G4EMLOW version: 8.6.1
• OS: Ubuntu 24
• Physics List: FTFP_BERT + G4EmLivermorePhysics
• Analysis: ROOT 6.x

Summary
The K-shell fluorescence yield for Silicon (Z=14) in G4EMLOW8.6.1 is 0.93%,
compared to the experimental reference value of 4.5% (Krause 1979),
representing a systematic underestimation by a factor of 4.8.

This affects simulations of Si detectors (SDD, Si(Li))

Geometry:
• Source: Eu-152 (14 keV gamma)
• Detector: Silicon cylinder (450 µm thick, 7.98 mm diameter)
• Simplified SDD geometry with Be window

Physics Configuration:
• ReplacePhysics(new G4EmLivermorePhysics)
• RegisterPhysics(new G4RadioactiveDecayPhysics)
• em->SetFluo(true)
• em->SetAuger(true)
• em->SetAugerCascade(true)
• em->SetDeexActiveRegion(“DefaultRegionForTheWorld”, true, true, true)
• em->SetDeexcitationIgnoreCut(true)
• Production cuts: gamma=0.01µm, e-=0.01µm

Log verification confirms:
✓ Fluorescence enabled: 1
✓ Auger electron cascade enabled: 1
✓ De-excitation module ignores cuts: 1
✓ Process “phot” using LivermorePhotoElectric with Fluo

Simulation Setup:
• 1000 photons at 13.95 keV incident on Si (450 µm)
• All fluorescence flags enabled and verified in logs
• PhysicsList correctly configured with SetDeexcitationIgnoreCut(true)

Results:
• 3015 photoelectric interactions (99.9% - correct)
• 28 K-alpha photons at 1720 eV created
• Observed yield: 28/3015 = 0.93%

G4EMLOW Analysis:
• File fl-tr-pr-14.dat sum: 0.93% ✓ (matches simulation)
• Literature (Krause M.O., J. Phys. Chem. Ref. Data 8, 307, 1979): 4.5%
• Discrepancy: Factor 4.8×

G4EMLOW Analysis:
• File fl-tr-pr-14.dat sum: 0.93% ✓ (matches simulation)
• Literature (Krause M.O., J. Phys. Chem. Ref. Data 8, 307, 1979): 4.5%
• Discrepancy: Factor 4.8×

Extended tosome other light elements
Similar underestimation for Z < 20:
• C (Z=6): 0.10% vs 0.27% (factor 2.7×)
• O (Z=8): 0.30% vs 0.76% (factor 2.5×)
• Al (Z=13): 1.50% vs 3.64% (factor 2.4×)
• Si (Z=14): 0.93% vs 4.50% (factor 4.8×)
• S (Z=16): 2.80% vs 6.59% (factor 2.4×)

Good agreement for Z > 25 (Fe, Cu, etc.)

Impact on my simulation
For Si detectors: escape peaks underestimated 5×

QUESTIONS

  1. Is this a known limitation of G4EMLOW for light elements?
  2. Are there plans to update with more complete fluorescence data?
  3. Should we apply correction factors for Si detector simulations?

Hello,

thank you very much for clear report. Your report is copied to the geant4 Bugzilla bug report system:

Please, visit and follow the evolution.

VI